web
Sie sind offline. Dies ist eine schreibgeschützte Version der Seite.
close
Zum Hauptinhalt wechseln
Tektronix, Inc.

  • TekTalk
        • All
        • Forums
  • Deutsch
    • Français
    • English
    • Deutsch
    • русский
    • 한국어
    • 日本語
    • 中文(台灣)
    • 中文(中国)
  • Login
  • Home
  • Profile

Tektronix Applications Expert

Tektronix Applications Expert

Beigetreten: 24.09.2022

  • Blog Posts
  • Ideas
  • Forum Posts
  • RE: Accessories needed for Calibration?

    Fri, 30 Sep 2022 19:51:25 GMT – Accessories needed for Calibration?

    There is no need to send the accessories. Please do not include them with your RMA.

  • RE: Semiconductor Sensor 4-Wire Measurement for DMM6500

    Fri, 30 Sep 2022 19:51:23 GMT – Semiconductor Sensor 4-Wire Measurement for DMM6500

    Application requires Dry Circuit Resistance capability. DMM6500 does not have this feature. If you are looking for this capability, the DMM75100 is capable.

  • RE: Self-Calibration/ Updating Internal Table

    Fri, 30 Sep 2022 19:51:21 GMT – Self-Calibration/ Updating Internal Table

    Thanks for contacting us,

    There is not a mechanism for self-calibration of the RSA306B. While you can follow the procedures in the Performance Verification manual (https://download.tek.com/manual/RSA306B-Real-Time-Spectrum-Analyzer-Specifications-Performance-Verification-077124301.pdf) to characterize the instrument and verify it is within spec, if the results indicate it needs to be recalibrated it will need to be sent to a calibration lab that supports it. You can find more detail on our calibration services webpage here: https://www.tek.com/en/services/calibration-services

  • RE: Seeing a "step" in the spectrum for the RSA306B

    Fri, 30 Sep 2022 19:51:19 GMT – Seeing a "step" in the spectrum for the RSA306B

    What you are seeing is the effect of a limitation of the RSA306B. The capture bandwidth is limited to 40MHz. This causes the RSA to capture several acquisitions and stitch them together, in your case DC to 30MHz, 30MHz to 70MHz, and 70MHz to 100MHz. This is normal behavior for the RSA, and unless you are taking measurements right at the noise floor it should not interfere with your measurements.

  • RE: SDS and UN 38.3 Test Results

    Fri, 30 Sep 2022 19:51:16 GMT – SDS and UN 38.3 Test Results

    Attached is the document from tek.com for the TDS3BATC.
    https://www.tek.com/en/msds-material-safety-data-sheets/lithium-ion-batteries-li-ion

    https://www.tek.com/en/search?keywords=TDS3BATC&facets=_templatename%3dMarketing+Document%26parsedtechnicaldocumenttype%3dMSDS+-+Material+Safety+Data+Sheets&sort=desc

  • RE: Issue with Screen Shutting Off for DPO3014

    Fri, 30 Sep 2022 19:51:13 GMT – Issue with Screen Shutting Off for DPO3014

    Your unit will need to be sent in for service. Please see below on how to proceed:


    To set up a repair RMA (model and serial number)
    call 1-800-833-9200 opt 2. Or 1- 800-438-8165
    Or send an email to rmarequest@tektronix.com
    Or https://www.tek.com/en/service-quote
    Or https://www.tek.com/en/services/repair/mytek_rma_request

  • RE: Screen Replacement for TDS2024

    Fri, 30 Sep 2022 19:51:11 GMT – Screen Replacement for TDS2024

    I'm not comfortable saying yes as there were multiple versions of the display and it's possible that there might be serial number breaks ect that could impact compatibility.

    That said the initial display module had the same part number when they were first introduced regardless of how many channels.

    https://www.tek.com/en/oscilloscope/tds1002-manual/tds1000-tds2000-series-service-manual

  • RE: Screen Issues with 4200

    Fri, 30 Sep 2022 19:51:08 GMT – Screen Issues with 4200

    Your unit will need to be sent in for service. Please see below on how to proceed:


    To set up a repair RMA (model and serial number)
    call 1-800-833-9200 opt 2. Or 1- 800-438-8165
    Or send an email to rmarequest@tektronix.com
    Or https://www.tek.com/en/service-quote
    Or https://www.tek.com/en/services/repair/mytek_rma_request

  • RE: Screen issues with TDS210

    Fri, 30 Sep 2022 19:51:05 GMT – Screen issues with TDS210

    The TDS210 is out of long term support and we no longer offer the displays. The old cost to replace the display actually was similar to what it would cost for a new scope. New scope would have multiple year warranty and include new probes.

    https://www.tek.com/en/products/oscilloscopes/tbs1000

    Let us know if you have any additional questions.

  • RE: Issue with screen on TDS2000

    Fri, 30 Sep 2022 19:51:04 GMT – Issue with screen on TDS2000

    Your TDS2000 will need to be sent in for service. Please see below on how to proceed:


    To set up a repair RMA (model and serial number)
    call 1-800-833-9200 opt 2. Or 1- 800-438-8165
    Or send an email to rmarequest@tektronix.com
    Or https://www.tek.com/en/service-quote
    Or https://www.tek.com/en/services/repair/mytek_rma_request

  • ‹
  • 1…
  • 11
  • 12
  • 13
  • 14
  • 15
  • 16
  • 17
  • 18
  • 19
  • …58
  • ›
This site is for Tektronix Customers to Learn, Share and Collaborate.
 

© 2025 TEKTRONIX, INC. | Terms of Use | Code of Conduct | Contact us

▼