• RE: Low filter function for MSO series

    Hi Hamed,<br> <br> If you wish to reduce noise on your signal then you can enable averaging or high-res acquisition mode.<br> <br> I will say that what I see from your image doesn't look like noise, it looks like ringing at the top of the rising edge, which is there for a reason and should not be ignored.
  • RE: Low filter function for MSO series

    Hi Hamed,<br> <br> We do not have an &quot;MSO series&quot;, the MSO terminology is used in several different series. Instead I'd ask for the full model number of your machine.<br> <br> If you have an MSO2/4/5/6, then you can go into your vertical channel menu and set the bandwidth limit as you desired, which applies a DSP low-pass filter to your channel input.<br> If you have the user defined filter option enabled, then you can also use that to generate more complex filters with the desired properties.
  • RE: HV Probe for THS710

    <span><span class="ui-provider cab cac c d e f g h i j k l m n o p q r s t cae caf w x y z ab ac ae af ag ah ai aj ak" dir=ltr>Hi,<br> <br> The <a aria-label="Link P5102" class="fui-Link ___1qmgydl f3rmtva f1ewtqcl fyind8e f1k6fduh f1w7gpdv fk6fouc fjoy568 figsok6 f1hu3pq6 f11qmguv f19f4twv f1tyq0we f1g0x7ka fhxju0i f1qch9an f1cnd47f fqv5qza f1vmzxwi f1o700av f13mvf36 f1cmlufx f9n3di6 f1ids18y f1tx3yz7 f1deo86v f1eh06m1 f1iescvh ftqa4ok f2hkw1w fhgqx19 f1olyrje f1p93eir f1h8hb77 f1x7u7e9 f10aw75t fsle3fq" href="https://w140.com/tekwiki/wiki/P5102" rel="noreferrer noopener" style="" target=_blank title="https://w140.com/tekwiki/wiki/p5102">P5102</a> is the highest voltage probe meant for the THS710.<br> <br> The P5122 maintains isolation, but the maximum input capacitance of the THS710 (27 pF) is out of the maximum compensation range of P5122 (22 pF), so you might not be able to compensate it properly.<br> <br> If isolation isn't a requirement (meaning the scope and probe aren't floated), then the P5100A and P6015A should work with the THS710.</span></span>
  • RE: Group Channels Via Scpi Commands

    Hi,<br> <br> The SCPI commands for querying and setting the channel grouping are still undocumented. I have reached out to our product line team to fix this.<br> <br> In the meanwhile, I have had success with <code>MAINWINDOW:SOURCEORDER</code>, but it's not perfect and I wouldn't recommend you use it yet.
  • RE: oscilloscope connector to current amplifier

    Hi,<br> <br> What you need is a TekProbe to TekVPI adapter: <a href="https://www.tek.com/en/sitewide-content/datasheets/t/p/a/tpa-bnc">https://www.tek.com/en/sitewide-content/datasheets/t/p/a/tpa-bnc</a>
  • RE: Tektronix MDO4104B-6 oscilloscope failing to respond during traces' collection process

    Hi,<br> <br> Adding sleeps is not the correct way to handle timeouts. VISA timeout errors should be caught and handled by the program; this usually means&#160; repeating the operation a few times and, if it still fails, exiting with an error.<br> Additionally, it's also important to use &quot;*ESR?&quot; to check for errors and &quot;*OPC?&quot; for synchronization.<br> <br> I have attached an example Python3 script for getting a curve query from an MSO4/5/6, but it will likely work (perhaps with some modifications) for your MDO4000B.
  • RE: DMM/Frequency Counter option for my MDO4104-6

    Hi Stan,<br> <br> I believe this was already answered in one of the TekScopes group, but I will answer here for others to see:<br> <br> The non-letter MDO4000 does not have DVM and frequency counter capabilities, so registering them does not unlock that option.<br> <br> For scopes that do feature that capability, you will receive an email with the license key after registering the instrument.
  • RE: Tektronix MDO4104B-6 oscilloscope failing to respond during traces' collection process

    Hi Birjit,<br> <br> I would ask how you're communicating with the instrument (USB, Ethernet, etc) and how you're handling VISA timeout errors?<br> It's normal to have timeouts, especially during the transfer of long waveforms or files, so it's important that your code handles these properly. It might be useful to post the relevant code.
  • RE: Ethernet Signal Mask on MSO44 with 4-SRENET

    Hi Gary,<br> <br> 4-SRENET is for Ethernet triggering and analysis (aka decoding), not compliance testing.<br> Our compliance testing package is a part of TekExpress, which is a Windows based application and so requires an oscilloscope running Windows. The MSO4 is not capable of running Windows, you will need an MSO5 or higher to use TekExpress.<br> <br> Please contact your account manager if you wish to return the instrument.
  • RE: AFG1022 and Rack Mount Kit

    Hi,<br> <br> We no longer sell a rack mount kit for the AFG1000.