  • RE: MSO4034B mixed signal oscilloscope compatibility with the TIVPWS500X probe

    Thu, 29 Sep 2022 23:17:09 GMT – MSO4034B mixed signal oscilloscope compatibility with the TIVPWS500X probe

    Hello,

    The current TIVP IsoVu Isolated Probes will only work with our newer MSO44, MSO46, MSO5 Series and MSO6 Series scopes. They will not work on the MSO4034B.
    There were older IsoVu Isolated Probes the TIVM and TIVH Series IsoVu™ probes that would work on the MSO4034B but they are out of production now.

  • RE: Need new P2220 oscilloscope probe tip-clips

    Thu, 29 Sep 2022 23:17:07 GMT – Need new P2220 oscilloscope probe tip-clips

    Hello,
    The manual for the P2220 is on Tek.com see link below.
    https://www.tek.com/en/passive-probe-manual/p2220-200-mhz-1x-10x-passive-probe
    The hook tip part number is 013-0107-08

    To order a part call 1-800-833-9200 opt 1.
    Or send an email to buy@tektronix.com

  • RE: MSO44 Minimum Voltage Range

    Thu, 29 Sep 2022 23:17:05 GMT – MSO44 Minimum Voltage Range

    Hello,

    Both of these are 10X attenuation probes, so the scope needs to multiply the voltage at the interface by 10. This is why you see a 5mV/Div minimum. To avoid this, you would need to use a 1X probe.

  • RE: MDO32 Minimum Record Length

    Thu, 29 Sep 2022 23:17:04 GMT – MDO32 Minimum Record Length

    1k points is minimum on MDO32

  • RE: Memory in the TDS3054B and declassification procedure

    Thu, 29 Sep 2022 23:17:02 GMT – Memory in the TDS3054B and declassification procedure

    Hello,

    Below you will find a link to the TDS3000B series declassification document. You can search for an instrument's declassification document by filtering the search type to Manual > Declassification.

    https://www.tek.com/en/oscilloscope/tds3014b-manual/tds3000-and-tds3000b-rev

  • RE: Measuring TBS1064 frequency in a single cycle

    Thu, 29 Sep 2022 23:17:01 GMT – Measuring TBS1064 frequency in a single cycle

    Hello,

    The frequency measurement algorithm displays the first complete cycle in the record. You can also use cursors to measure the exact cycle you want measured.

  • RE: SMU236 Spurs across 200MHz Bandwidth

    Thu, 29 Sep 2022 23:16:59 GMT – SMU236 Spurs across 200MHz Bandwidth

    Using an SMU236 and an Agilent N9010A spectrum analyzer measuring the voltage difference across a discrete 24K resistor supplying a 20uA current, the recommendation is to apply a Low-Pass Filter on the SMU Output

  • RE: AFG3011C Sini Wave Frequency

    Thu, 29 Sep 2022 23:16:58 GMT – AFG3011C Sini Wave Frequency

    Yes the AFG3011C can produce a 10MHz 20Vpp continuous sine wave into 50Ω.

  • RE: KICKSTARTFL-SUITE Free Trial

    Thu, 29 Sep 2022 23:16:56 GMT – KICKSTARTFL-SUITE Free Trial

    Thank you for evaluating Kickstart for your photovoltaic cell testing. Unfortuntately trial extensions for Kickstart are not available.

  • RE: Floating License on DPO72004C

    Thu, 29 Sep 2022 23:16:55 GMT – Floating License on DPO72004C

    Please check the time in Windows on the scope is set correctly.
    Please check the clock on the scope UI to see if it matches the time in Windows on the scope.
    1. Set the time in Windows so that it is correct.
    2. In the scope go to Utilities -> Set Time and Date…
    3. In the Set Time and Date Pane click the “Get date and time” button then click the “Set time and date now” button.
    4. Restart the scope

  • ‹
  • 1…
  • 28
  • 29
  • 30
  • 31
  • 32
  • 33
  • 34
  • 35
  • 36
  • …58
  • ›
This site is for Tektronix Customers to Learn, Share and Collaborate.
 

© 2025 TEKTRONIX, INC. | Terms of Use | Code of Conduct | Contact us

▼