Thanks for the verification. Since I was trying to record 25us @ 100MSps maybe my horizontal div setting was what was limiting me to the 50MSps and 500MSps selections on the screen. When I get my scope back from a coworker I'll check for combinations that provide 100MSps like your picture.
-
Thank you for the reply. Will I get more choices through SCPI for the sample rate than I seem to get through the front panel?<br>
<br>
For example I want 100MSps but through the front panel I only seem to be able to select record lengths that give me 50Msps or 500Msps.
-
I'm sorry if this is a stupid question but I'm used to this being easy and I can't figure it out with this scope.<br>
<br>
I'm reading waveforms from an Tektronix MDO34 that do need the full 2.5GSps sampling rate. The ethernet communication is very slow if I read the entire waveform at that rate so I'd prefer to read every 25th sample to get it at 100MSps. With other equipment I have a divisor or sparsing factor that I can set to make this easy.<br>
<br>
From the front panel I can see setting the record length affects the sample rate although it seems the finite number of choices does not give me 100MSps. I have tried using ":hor:reco <num_points>" and ":wfmi:xin <sample_period>" to have the desired effect but what I read is not my waveform.<br>
<br>
Can anyone help me figure out how to achieve this task with this (and I assume all) Tektronix scopes?