  • RE: TBS2000B/MSO2024 AC Signal Measurement

    Thu, 29 Sep 2022 23:10:49 GMT – TBS2000B/MSO2024 AC Signal Measurement

    Hello,

    AC coupling is not available for triggering on those two models. Low Frequency Reject would be the closest option with these two scopes. The cutoff frequency for LF Rejection is 65kHz.

  • RE: MDO4034B-3 oscilloscope AC gain accuracy

    Thu, 29 Sep 2022 23:10:47 GMT – MDO4034B-3 oscilloscope AC gain accuracy

    Hello,

    Unfortunately, we do not specify the AC gain accuracy for this scope. Our full list of specifications is in the Performance Verification manual: https://www.tek.com/en/oscilloscope/mso4000-dpo4000-manual/mso4000b-and-dpo4000b-series-0.

  • RE: TBS1032B 9V Circuit - Stopped Working

    Thu, 29 Sep 2022 23:10:45 GMT – TBS1032B 9V Circuit - Stopped Working

    Hi,

    Please power off, and remove power then power back on and do a default setup to verify that the the unit passes the Self Calibration. From the Utility button "Do Self Cal".

  • RE: Need to generate 9 V@100A DC

    Thu, 29 Sep 2022 23:10:44 GMT – Need to generate 9 V@100A DC

    Hi, I recommended the 2260B-30-108

  • RE: 650-6122-00 probe tips repair

    Thu, 29 Sep 2022 23:10:43 GMT – 650-6122-00 probe tips repair

    Hello ,

    The two parts that you list are probe tips for our IsoVu probes. Neither of the probe tips are repairable. We offer the tips as repalceable accessory for the IsoVu probes. 650-6122-00 is actually ordered as TIVPMX10X. 650-6192-00 is actually ordered as TIVPWS500X

    TIVPMX10X
    MMCX 10X Tip Cable (SHORT)
    $1090.00 ea. USD

    TIVPWS500X
    WSQPIN 500X Tip Cable (SHORT)
    $1090.00 ea USD

    If you are interested in placing order or parts pricing, please reach out to our sales at 1.800.833.9200 option 1 or email the request to buy@tektronix.com

  • RE: 600V input?

    Thu, 29 Sep 2022 23:10:41 GMT – 600V input?

    Hello,

    We do offer several probes that would allow for you to safely measure up to 600V with the TBS1052C. Below you can find links to both our high voltage single ended and high voltage differential probes. For your use case consider either the P5200A for differential or the P5122 for single ended measurements.

    https://www.tek.com/en/products/oscilloscopes/probes/high-voltage-differential-probes

    https://www.tek.com/en/products/oscilloscopes/probes/high-voltage-probe-single-ended

  • RE: TDP0500 500 MHz High Voltage Differential Probes not working

    Thu, 29 Sep 2022 23:10:40 GMT – TDP0500 500 MHz High Voltage Differential Probes not working

    Hello,
    If this probe or others like have past before then the unit should come in for a repair. If not you can send in the TDP0500 to our Factory service team can verify and repair if needed.

    To set up a repair RMA (model and serial number)
    call 1-800-833-9200 opt 2.
    Or send an email to rmarequest@tektronix.com
    https://www.tek.com/en/services/repair/mytek_rma_request

  • RE: Seeing 40V limit on 2450

    Thu, 29 Sep 2022 23:10:38 GMT – Seeing 40V limit on 2450

    This would likely be due to the interlock not being engaged. Please refer to the 2450 manual: https://www.tek.com/en/source-measure-units/keithley-smu-2400-series-sourcemeter-manual/models-2450ec-2460ec-and-2461ec-potentiostats-and-galvanostats-users-manual

  • RE: 4 wire resistance measurement with Keithley 2000, Keithley 2400, or Keithley 2812?

    Thu, 29 Sep 2022 23:10:36 GMT – 4 wire resistance measurement with Keithley 2000, Keithley 2400, or Keithley 2812?

    These requirements would be pushing the capabilities of the 2000 DMM. Therefore, I recommended you use the 2400 and 2182A combination.

  • RE: 4 wire resistance measurements with 2000 DMM?

    Thu, 29 Sep 2022 23:10:35 GMT – 4 wire resistance measurements with 2000 DMM?

    For milliohm measurements the Keithley 2000 would use 100-ohm range, with a 1mA source current.

  • ‹