web
您已離線。這是該頁面的唯讀版本。
close
跳至主要內容
Tektronix, Inc.

  • TekTalk
        • All
        • 論壇
  • 中文(台灣)
    • Français
    • English
    • Deutsch
    • русский
    • 한국어
    • 日本語
    • 中文(台灣)
    • 中文(中国)
  • Login
  • Home
  • Profile

Tektronix Applications Expert

Tektronix Applications Expert

加入日期/時間: 2022/9/24

  • Blog Posts
  • Ideas
  • Forum Posts
  • RE: Differential Probe for TBS2102B

    Thu, 29 Sep 2022 23:14:00 GMT – Differential Probe for TBS2102B

    The P5200A uses a standard BNC interface and is compatible with the TBS2102B. However you will have to manually set the attenuation on the scope. You can find information on how to adjust the channel parameters in the TBS2102B user manual.

  • RE: Part number for TPP0250 Witches Hat

    Thu, 29 Sep 2022 23:13:58 GMT – Part number for TPP0250 Witches Hat

    The part number is 013-0362-00 we have stock the cost is about $15. Note that Part number 013-0362-xx Hook Tip has been replaced with Part number 013036201. The 013036201 was improved to reduce fall out in MFG.

    Manual link below if you need other parts.
    https://www.tek.com/en/passive-probe-manual/tpp0250-tpp0500b-1

    To order a part call 1-800-833-9200 opt 1.
    Or send an email to buy@tektronix.com

  • RE: Firmware for TDS3014C

    Thu, 29 Sep 2022 23:13:57 GMT – Firmware for TDS3014C

    Please try to reinstall the firmware using the installation instructions found with the firmware. Also add an empty file titled "forceinstall.txt" to the root folder this will force the scope to install the firmware. This is also useful should you need too downgrade or reload the scopes firmware in the future.
    Latest firmware for the TDS3000C:
    https://www.tek.com/en/support/software/firmware/firmware-upgrade-tds3000c-series-oscilloscopes-v431

  • RE: Calibrating the MSO4104B

    Thu, 29 Sep 2022 23:13:55 GMT – Calibrating the MSO4104B

    The Service manual for the product has the information for the adjustment procedure, Service manual part number 077051202
    https://www.tek.com/en/oscilloscope/mso4000-dpo4000-manual/mso4000b-and-dpo4000b-series-1
    The procedure requires a digital probe calibrator which puts the same signal on all inputs at the same time. The procedure is an automated procedure in the unit and you must start at the start and go all the way through to the end with no errors or failures, Which could be many many hours.
    Double check different probe?
    If the Digital input needs to be calibrated you can send the unit into our factory for calibration.

    To set up a repair RMA (model and serial number)
    call 1-800-833-9200 opt 2. Or 1- 800-438-8165
    Or send an email to rmarequest@tektronix.com
    Or https://www.tek.com/en/service-quote
    Or https://www.tek.com/en/services/repair/mytek_rma_request

  • RE: High Current DC Probe

    Thu, 29 Sep 2022 23:13:53 GMT – High Current DC Probe

    We do not have a DC current probe that meets your 600A requirement. The closest we offer would be the TCP404XL. The TCP404XL is rated for 500A continuous or 750A non-continuous. The main constraint at these current levels is heat, as you mentioned. If depending on ambient temperature or length of your test, this probe might work OK for you at 600A. The specifications pertaining to this start on PDF p. 76 in the User Manual: https://www.tek.com/en/current-probe/current-probe-manual/tcpa300400--tcp300a400-series

    As you can see from PDF p. 77, at an ambient temperature of 23 degrees C you can measure 600A continuously. However, there's some restrictions and warnings in this section (starting on p. 77) that I would recommend you cover thoroughly to make sure this will work for your test conditions. 600A continuous can be done but is right near the limit for this probe. Here's another helpful link: https://www.tek.com/en/datasheet/ac-dc-current-measurement-systems

  • RE: MSU of MSO5000B

    Thu, 29 Sep 2022 23:13:52 GMT – MSU of MSO5000B

    Tektronix scopes do not specify an MSU (Measurement System Uncertainty) specification, but we do offer an ISO Guide to the Expression of Uncertainty in Measurement (GUM) with our Cal Certs. The coverage factor (k) is 2, with coverage probability of approximately 95 %. Uncertainty values are reported with the same units as the measured values. Statements of conformity to specifications account for uncertainty. Statements of conformity to specifications are made with less than 2% global false acceptance risk for each measurement result.

    This information is specified directly from our Calibration certifications (Cal Cert). You should have received this calibration document when you first received the instrument.

    If you want to reference our document that contains the specifications for the MSO5000B, I have linked the Performance verification manual below.
    MSO5000B Specifications and Performance Verification: https://www.tek.com/en/oscilloscope/dpo70000-mso70000-manual/mso-dpo70000dx-mso-dpo70000c-dpo7000-c-mso5000-b-and-dpo5000
    • MSO5000B Specs are on page 77 (PDF page 89)

    I hope this helps!

  • RE: Need to purchase 016-2016-xx

    Thu, 29 Sep 2022 23:13:50 GMT – Need to purchase 016-2016-xx

    We have two version of this part number. See below:
    1) The 016-2016-00 is intended for use by existing customers that have previously purchased and designed the -00 version into their circuit board. 2) The 016-2016-01 is a newer version that allows the user more flexibility when designing it into circuit board. The -01 is not a direct replacement for the -00 version.

    To order a part call 1-800-833-9200 opt 1.
    Or send an email to buy@tektronix.com

  • RE: Soft case for P643 probe

    Thu, 29 Sep 2022 23:13:49 GMT – Soft case for P643 probe

    The soft case 016195203 will work with your P6243. Please email buy@tektronix.com to request a quote for the case.

  • RE: TBS2204B - Isolated Power?

    Thu, 29 Sep 2022 23:13:47 GMT – TBS2204B - Isolated Power?

    The TBS2204B does not have isolated power. It is a grounded scope.

  • RE: PC Software for TDS2012

    Thu, 29 Sep 2022 23:13:46 GMT – PC Software for TDS2012

    TekScope Utility will work with your TDS2012 and supports RS-232 and GPIB. This utility will allow you to pull and save data from your scope. Let me know if you have questions about the software or if it does not meet your needs. Learn about TekScope here: https://www.tek.com/en/products/software/tekscope-pc-analysis-software

  • ‹
  • 1…
  • 40
  • 41
  • 42
  • 43
  • 44
  • 45
  • 46
  • 47
  • 48
  • …58
  • ›
This site is for Tektronix Customers to Learn, Share and Collaborate.
 

© 2025 TEKTRONIX, INC. | Terms of Use | Code of Conduct | Contact us

▼